Best way to implement serial communication - gill windmaster


FedeJack Oct 10, 2025 09:41 PM

Hi, I want to connect Gill Windmaster via Serial (RS232) and collect data every 50 msec. The bound rate at 9600 is limited by the length of the cable, which is 6.5m. Here the working code I used:

 

'CR6 Series & Gill WindMaster RS232
Const STX=CHR(2) 'ASCII start-of-text
Const ETX=CHR(3) 'ASCII end-of-text
Public row1 As String * 50
Public istart1 As Long
Public tmp As String * 50

'Declare Public Variables
Public u_1
Public v_1
Public w_1
Public Ts_1

'Define Data Tables
DataTable (Sonic,True,-1)
  DataInterval(0,50,mSec,0)
  TableFile ("CRD:"&Status.SerialNumber(1,1)&"_sonic_data_",64,-1,0,1,Day,0,0)
  Sample(1, u_1,FP2)
  Sample(1, v_1,FP2)
  Sample(1, w_1,FP2)
  Sample(1, Ts_1,FP2)
EndTable

'String out:"Q,+000.00,-000.01,+000.02,M,+022.55,00,"
BeginProg
	SerialOpen(ComC1,9600,3,0,512,0)
	Scan(50,mSec,1,0)
		SerialIn(row1,ComC1,0,ETX,50) 'read until ETX > complete string ()
		istart1 = InStr(1,row1,"Q",2)
		tmp = Mid(row1,istart1,40)
		If (istart1 = 0) OR (Len(tmp) <> 40) Then
		  SerialFlush (ComC1)
		Else
		  u_1 = Mid(row1,istart1+2,7)
		  v_1 = Mid(row1,istart1+10,7)
		  w_1 = Mid(row1,istart1+18,7)
		  Ts_1 = Mid(row1,istart1+28,7)
		EndIf
  	
		CallTable Sonic
	NextScan
EndProg


 The code works but I had to introduce the tmp and if statement to exclude cases where the saved string was truncated, with the middle characters missing like

 

Q,+022.55,00,

 These cases appear once every 10 min if I remove the "If (istart1 = 0) OR (Len(tmp) <> 40)" part.

My question:

1) There is a best way to implement a fast serial read communication?

2) If I replace SerialIn() with

 

SerialInRecord (ComC1,row1,"Q",0,ETX,n1,01)

 

no rows are stored, why?

Thanks


JDavis Oct 13, 2025 02:32 PM

If using SerialInRecord(), remove SerialFlush() from your program. Use the number of bytes returned by SerialInRecord to indicate if there is new data to process.


FedeJack Oct 14, 2025 12:45 PM

Thanks, so both SerialIn() and SerialInRecord() codes are equally efficient or one of the two is preferable?


JDavis Oct 14, 2025 03:33 PM

SerialInRecord() will be more reliable.

I suggest looking at the example program Campbell Scientific provides for the Windsonic1 (RS-232). The data is a very similar format.

FYI, I think at 9600 baud and the ASCII format, you won't be able to achieve 50ms output. It takes a little over 1 ms per character to be transmitted at that baud rate. You can probably do RS-232 at 19200 baud with that cable length. If there is too much electrical noise at site to run that baud rate, change the sensor and CR6 to RS-485 half duplex. 