Capturing Serial Data


Al_Water Feb 16, 2017 05:29 PM

Hi, 

We recently had to buy a new upgraded temperature sensor that we utilize to monitor a water bath (a VWR digital thermometer).  The sensor outputs a consistent data stream via RS232.  We used a CR800 to capture the data for different time tables utilizing the RS232 com.  Although the new sensor seems to output the same data stream as the old sensor I cannot capture or see the data on the CR800.  Both sensor's output data looks similar when viewing through a hyper terminal Program:

C,26.80

C,26.80

C,26.80

C,26.80

C,26.79 – New Sensor

 

C, 24.15 ,

C, 24.15 ,

C, 24.15 ,

C, 24.15 ,

C, 24.15 , - Old Sensor

I wrote the following program to capture the old sensor data:

BeginProg
Scan (5,Sec,0,0)
SerialOpen (ComRS232,9600,0,0,16)
SerialInRecord (ComRS232,VWRTemp,67,8,0,NBytesReturned,01)
StringTemp = VWRTemp
CorrTemp = Mid (StringTemp,2,6)
CallTable Temp5Sec
NextScan
EndProg

I do not understand why the logger cannot see the data from the new sensor stating a "NAN" value.  I have played around with the scan interval, settings on the SerialOpen commend, beginword, endword, bytes size, string size and different commends.  Very frustrating!

Any help or suggestions would be greatly appreciated! Thanks. 


JDavis Feb 16, 2017 05:31 PM

In place of MID, try using SplitStr with the Numeric option.


Al_Water Feb 16, 2017 05:40 PM

Would that change the "NAN" value on the VWRTemp variable?  Should I not be seeing the data there first?


JDavis Feb 17, 2017 11:53 PM

Your new sensor outputs fewer bytes. Change the 8 to a 6 in SerialInRecord.